Alarm conditions |
English:PDF | German:PDF |
Table of Contents
- 1 Functional description
- 2 Alarm conditions in the evaluation strategy
- 3 Alarms for variablecharacteristics
- 3.1 Single value outside n% of a two-sided tolerance
- 3.2 Warning limit from n% of a two-sided tolerance
- 3.3 Single value outside n% of a tolerance with naturallimits
- 3.4 Warning limit from n% of a tolerance with natural limit
- 3.5 Single value outside a one-sided tolerance
- 3.6 Acceptance limits violated
- 3.7 Single value outside the scrap limits
- 3.8 Control limit violation in location chart
- 3.9 Alarm limit violation in locationchart
- 3.10 Warning limit violation in locationchart
- 3.11 Specification limit violation location chart
- 3.12 Run, trend or middle thirdin location chart
- 3.13 Variation below control limit
- 3.14 Variation above control limit
- 3.15 Variation below warning limit
- 3.16 Variation above warning limit
- 3.17 Run, trend, middle thirdin variation chart
- 3.18 Requirements according to evaluation strategy
- 3.19 Mean value outside n% of a tolerance with trend process
- 3.20 Warning limit from n% of a tolerance with trend process
- 3.21 Single value eliminated by strategy
- 3.22 Western Electric Rules LocationChart
- 3.23 Western Electric Rules Variation Chart
- 3.24 Implausible measured value (attribute 290)
- 4 Alarms for attributive characteristics
- 5 Common use of alarms
1 Functional description
Based on the alarm conditions, the evaluation strategy used in (O-QIS) procella defines when a measured value is considered an alarm and is alerted/displayed accordingly.
This document shows all available alarm conditions by means of examples. (Version 13.0.5.1)
|
NOTE: |
In this document only the available alarm conditions are shown and described. The general requirements for alarm generation are not explained further. |
|
NOTE: |
In order to generate the appropriate alarms in the procella application, it is mandatory to consider the interaction of different options. · Create and save a corresponding QCC · Consideration of the deposited/stored QCC based on the evaluation strategy used. Further information on the topic "Working with Quality Control Charts" can be found in a separate, topic-related documentation. |
2 Alarm conditions in the evaluation strategy
The alarm conditions for variable and attributive characteristics can be selected via "O-QIS" within the evaluation strategy.
It is possible to set alarms for variable (K2004=0) and attributive (K2004=1) characteristics.
It is also possible to specify an explicit default for individual characteristics based on the characteristic class (K2005).
The time at which the alarm dialog should be displayed is described in detail in the document "procella configuration data acquisition" under "Alarm dialog".
3 Alarms for variablecharacteristics
The following alarm conditions are available for variable characteristics (K2004 = 0):
|
NOTE: |
All examples listed here use the option "Center around tolerance middle" for two-sided tolerances. |
3.1 Single value outside n% of a two-sided tolerance
Here, a percentage input is used to specify when a measured value is considered an alarm.
In this example, the tolerance is a total of 0.2mm (100%) and the tolerance center is 10mm.
The detected measured value (10.12mm) is above the 100% of the tolerance specification and causes a corresponding alarm message:
3.2 Warning limit from n% of a two-sided tolerance
Here, a percentage specification is used to indicate when a measured value is considered an alarm. In this example, the alarm warning limit is 50% of the specified total tolerance (100% = 0.2 mm).
The detected measured value (10.07 mm) is above the specified tolerance warning limit and causes a corresponding warning message:
|
NOTE: |
The percentage specification of a warning limit only works if a tolerance is also specified. (Chap. 3.1) |
3.3 Single value outside n% of a tolerance with naturallimits
These are characteristics with natural boundaries (characteristics limited on one side) such as the roundness of a shaft or the flatness of a surface.
The tolerance here is 10.1 mm. (100%) and the natural limit is set at the lower specification limit.
If a measured value (e.g., 10,11) is detected which is above the specification of the natural limit (here OSG), a corresponding alarm message is output.
3.4 Warning limit from n% of a tolerance with natural limit
Here, a percentage input is used to specify when a measured value is considered an alarm.
In this example, the alarm warning limit is 50% of the specified total tolerance (100% = 7 mm).
As soon as a measured value (4.5 mm) above the specified tolerance warning limit is detected, this causes a corresponding alarm.
3.5 Single value outside a one-sided tolerance
This option is used to define that an alarm is displayed as soon as the existing tolerance limit is violated.
This option can only be activated/deactivated. A percentage setting is not possible here.
Here, for example, only an upper specification limit exists.
Now, as soon as a measured value above the upper specification limit is detected, a corresponding alarm appears.
3.6 Acceptance limits violated
This option is used to specify that an alarm is to be displayed as soon as the acceptance limits (K2116/K2117) stored for the respective characteristic have been violated.
This option can only be activated/deactivated. A percentage specification is not possible here.
If a measured value (here e.g., 9.42) is recorded below/above the acceptance limits, the corresponding alarm is displayed.
3.7 Single value outside the scrap limits
This option is used to specify that an alarm is to be displayed as soon as the scrap limits (K2114/K2115) stored for the respective characteristic have been violated.
This option can only be activated/deactivated. A percentage specification is not possible here.
If a measured value (here e.g., 10.71) is recorded below/above the acceptance limits, the corresponding alarm is displayed.
3.8 Control limit violation in location chart
This option is used to specify that an alarm is to be displayed as soon as the control limits stored / calculated for the respective characteristic have been violated within the location chart.
This option can only be activated / deactivated via the alarms.
The prerequisite for this is the previous saving of a corresponding QCC per characteristic.
The corresponding alarm is displayed as soon as the subgroup has been completely recorded. (Here, for example, a subgroup size of 3 is used).
If measured values (here e.g., 10.55 / 10.65 / 10.75) are recorded below/above the control limit in the location chart, the corresponding alarm dialog is displayed because
the subgroup statistic "Average" lies outside the upper control limit.
3.9 Alarm limit violation in locationchart
This option is used to specify that an alarm is to be displayed as soon as the alarm limits stored for the respective characteristic within the location chart (K8006/K8007) have been violated.
This option can only be activated/deactivated.
|
NOTE: |
The manually stored alarm limits must correspond to the control limits of the saved QCC. |
If measured values (here e.g., 10.55 / 10.65 / 10.75) are recorded below/above the alarm limit in the location chart, the corresponding alarm is displayed because the subgroup statistic "average " is above the upper alarm limit.
3.10 Warning limit violation in locationchart
This option is used to specify that an alarm is to be displayed as soon as the warning limits stored for the respective characteristic within the location chart (K8014/K8015) have been violated.
This option can only be activated/deactivated.
|
NOTE: |
The manually stored warning limits must correspond to the control limits of the saved QCC. |
If measured values (here e.g., 10.35 / 10.45 / 10.55) are recorded below/above the alarm limit in the location chart, the corresponding alarm dialogue will be displayed, because the subgroup statistics "average" is above the upper warning limit.
3.11 Specification limit violation location chart
This option is used to specify that an alarm should be displayed as soon as the mean/median (depending on the evaluation strategy) of a sample exceeds or falls below the specified characteristic specification limits (OSG/USG) within the location chart.
If measured values (here e.g., 9.65 / 9.55 / 9.45) are recorded below/above the specification limits in the location chart, the corresponding alarm dialogue will be displayed because the subgroup statistics "average" is outside the lower specification limit.
3.12 Run, trend or middle thirdin location chart
This option is used to specify that an alarm should be displayed as soon as a process within the location chart fulfils the specifications (run/trend/middle third) in the evaluation strategy used.
The prerequisite for this is the previous saving of a corresponding QCC per characteristic.
3.13 Variation below control limit
This option is used to specify that an alarm is to be displayed as soon as the lower control limit saved for the respective characteristic has been violated within the variation chart.
The prerequisite for this is the previous saving of a corresponding QCC per characteristic.
If measured values are now recorded (10.0 / 10.0 / 10.0) that provide a standard deviation of 0 within the variation charts, the corresponding alarm appears.
3.14 Variation above control limit
This option is used to specify that an alarm is to be displayed as soon as the upper control limit stored for the respective characteristic has been violated within the variation chart.
The prerequisite for this is the previous saving of a corresponding QCC per characteristic.
If measured values are now recorded (10.3 / 11.3 / 12.3) that cause a standard deviation of > UCL within the variation charts, the corresponding alarm dialogue appears.
3.15 Variation below warning limit
This option is used to specify that an alarm is to be displayed as soon as the lower warning limit stored for the respective characteristic has been violated within the variation charts.
The prerequisite for this is the previous saving of a corresponding QCC per characteristic.
If measured values are now recorded (10.3 / 10.3 / 10.3) that cause a standard deviation below the warning limit within the variation charts, the corresponding alarm appears.
3.16 Variation above warning limit
This option is used to specify that an alarm is to be displayed as soon as the upper warning limit stored for the respective characteristic has been violated within the variation charts.
The prerequisite for this is the previous saving of a corresponding QCC per characteristic.
If measured values are now recorded (10.5 / 10.8 / 11.1) that are within the variation charts for a standard deviation above the warning limit, the corresponding alarm dialogue appears.
3.17 Run, trend, middle thirdin variation chart
This option is used to specify that an alarm is to be displayed as soon as a process within the variation chart meets the specifications (run/trend/middle third) in the evaluation strategy used.
The prerequisite for this is the previous saving of a corresponding QCC per characteristic.
3.18 Requirements according to evaluation strategy
Please note that this alarm condition is notavailable in procella / O-QIS procella. This is an alarm condition that is only available in the Reporting System (M-QIS).
3.19 Mean value outside n% of a tolerance with trend process
3.20 Warning limit from n% of a tolerance with trend process
3.21 Single value eliminated by strategy
In this way, measured values that are selected by the corresponding specification in the evaluation strategy are automatically identified as outliers and removed.
Which default is to apply here is defined in the "Outliers" area of the evaluation strategy used. The measured values identified as outliers are not taken into account for the calculation of capability parameters.
Here, for example, a measured value (12.2) is automatically identified as an outlier on the basis of the plausibility limits (K2130/K2131) and removed.
The outlier is displayed as a dashed line.
3.22 Western Electric Rules LocationChart
This option is used to specify that an alarm is to be displayed as soon as the rule stored for the respective characteristic takes effect within the location chart.
This option can only be activated / deactivated via the O-QIS alarms.
The percentage default and/or a fixed value of the control limits incl. the rule to be used must be via the saved QCC.
The prerequisite for this is the previous saving of a corresponding QCC per characteristic.
3.23 Western Electric Rules Variation Chart
This option is used to specify that an alarm is to be displayed as soon as the rule stored for the respective characteristic takes effect within the variation chart.
This option can only be activated / deactivated via the O-QIS alarms.
The percentage default and/or a fixed value of the control limits incl. the rule to be used must be via the saved QCC.
The prerequisite for this is the previous saving of a corresponding QCC per characteristic.
3.24 Implausible measured value (attribute 290)
Please note that this alarm condition only applies in the CMM Reporting module within O-QIS. Depending on the CMM configuration, for example, the event dialog is displayed after confirmation (green check mark) in order to store the event/measure/cause. Alternatively, the alarm can also be displayed in the alarm viewer. This alarm option is only valid for variable characteristics.
A prerequisite for the alarm is that a measured value with attribute 290 is present within the uploaded data set.
|
NOTE: |
This alarm requires that the option "Take over incomplete sample in evaluation" (Preparation >> Outlier) is set in the evaluation rate used. If this is not the case, the corresponding measured value is deleted, and no alarm is generated. |
4 Alarms for attributive characteristics
For attributive characteristics (K2004 = 1), it is possible to specify a defect percentage or via parts per million (ppm):
The following alarm conditions are available for attributive characteristics:
4.1 Error percentage in sample
Based on a percentage specification of the error share, a corresponding alarm is displayed here during the recording of attributive characteristics.
The alarm is triggered as soon as the specified error percentage is exceeded.
For a sample size of 100, for example, the error percentage is 23. Accordingly, an alarm is displayed after the error percentage is entered.
4.2 Item rated as poor
Based on this alarm specification for a faulty attributive characteristic (niO/nOK), this alarm takes effect during acquisition.
As soon as a niO/nOK measured value is acquired, the corresponding alarm dialog appears.
4.3 Intervention boundary violation in discrete QCC
This option is used to specify that an alarm is to be displayed as soon as the control limits stored / calculated for the respective characteristic have been violated within the location chart.
This option can only be activated / deactivated via the alarms.
The percentage default and/or a fixed value must be defined via the saved QCC.
The prerequisite for this is the previous saving of a corresponding QCC per characteristic.
4.4 Implausible measured value (attribute 290)
Please note that this alarm condition only applies in the CMM Reporting module within O-QIS. Depending on the CMM configuration, for example, the event dialog is displayed after confirmation (green check mark) in order to store the event/measure/cause. Alternatively, an alarm can also be displayed in the alarm viewer. This alarm option is only valid for attributive, ordinal and nominal characteristics.
A prerequisite for the alarm is that a measured value with attribute 290 is present within the uploaded data set.
5 Common use of alarms
Common use of alarm conditions is generally possible. (One measured value generates several alarms).
For the display of the alarms in the alarm dialog, the stored alarm ranking is leading.
|
In the following example, two alarms are to be displayed based on a measured value that meets the following criteria: - Measured value must be outside a two-sided tolerance (50%outside) - Measured value violates the stored acceptance limit in K2116 / K2117
|