We have a problem with the machine. When calibrating the sensors, the program stops at the CalcTool command. Can anyone help?
Your Products have been synced, click here to refresh
We have a problem with the machine. When calibrating the sensors, the program stops at the CalcTool command. Can anyone help?
Hello Piotr, Can you let us have the set up (machine, sensor etc.) As well as the QualifyTool command line?
Dea Global 091508
PH10M, SP600M
Please do a weight balance, by turning the sensor on & off on the jog box.
Could you explain this to me in more detail?
Press the button top left on & off
I done that
I have found this to be common:
1) With really flimsy stylus systems
2) When foreign material has accumulated on stylus tip
3) When I accidentally build probe with wrong tip (like a 0.4mm, rather than a 0.3mm)