Trying to measure a cylinder using the rotary table and to create elements with the function CadGenCylCirs and scanning.
The machine behaves funny both the rotary table and the X and Y axes move at the same time.
The probe is at the same position on the part while the rotary table moves, following the rotary table around.
I have found a solution to the problem and that is to change the "Tolerances for Scanning types" in the point NCI in the NPT list from 1 to lower than 1,
then it works as it should, the rotary table rotates and the X and Y axes are still.
Is there somewhere where the parameter default value for "Tolerances for Scanning types" for CadGenCylCirs is stored, so I can put it in LDB or in GDB and avoid having to change the value every time I create new elements.
If you forget to mask the CadGenCylCirs lines in the program, the value changes back to 1 and the machine crashes
In the IPPSimulator it works even if the value is 1 on "Tolerances for Scanning types" so it's only a problem in the CMM.
Regards Anders C